Re: Digital Osci and Logic Analyzer



On Fri, 07 Apr 2006 15:08:34 GMT, nico@xxxxxxxxxxx (Nico Coesel) wrote:

Rich Webb <bbew.ar@xxxxxxxxxxxxxxxxxx> wrote:

As a rule of thumb, estimate the useable bandwidth as being limited by
one-tenth of the ADC sample rate, not one-half. That lets you recover
(more or less) the fifth harmonic of the signal of interest right at
Nyquist. For some commercial examples, there's the Fluke 199C: 2.5 Gsps,
overall bandwidth 200 MHz; and Tek TDS2012: 1 Gsps and 100 MHz. So for
your

I disagree here. Shannon says a signal can be reconstructed up to half
the sampling frequency. The higher sampling rates of the examples you
mention are probably used to use less bits for sampling and make the
analog input filter easier to construct (an RC filter still won't do
the job) to prevent frequency aliasing. Digital filtering does the
rest.

Thought problem. Signal A is a 1 MHz sine wave, signal B is a 1 MHz
square wave. Sample each at 2 Msps. Based only on the samples, and not
on any a priori knowledge of the expected signal characteristics, what
would the o'scope display at 1 usec/div'n?

Lets suppose I decide to buy a 1Gsps ADC(which isn't going to happen any
time soon) and I have a probe that works properly with the ADC. How would I
go about storing/streaming all those samples? This would require a memory
chip be able to work down at 1ns or so? I was thinking I could use several
gigs of pc memory in parallel to reduce the latency and increase total
sample size to a few seconds... is this possible? What about encoding for a
digital stream? do something like rle on the bit stream where 01 one would
store 2 bits as one and its number of repetitions? (need to use 2 bits
atleast so things like a clock signal are encoded efficiently)

You don't need to capture or store more samples than you need for one
display screen.

Assume that your display area is 500 pixels x 500 pixels at ten
divisions each for horizontal and vertical. No matter how fast you
sample, you've only that area to work with. For each time step (vertical
column) you have several choices.

I disagree again, the less memory you have the more difficult it will
be to capture a problem. You'll need some pretty fancy triggering.

Ultimately, however many samples are collected at whatever resolution
and at whatever rate, they're going to end up on a relatively low-res
display (low-res at least as compared to an "infinite" resolution analog
display). The 500x500 is just an example to make the numbers easy; most
are probably based on 1/4 VGA, VGA, or XGA resolutions.

You can decimate your sample rate down to where you grab just one
instantaneous sample at that tick and store/plot only that value. For
example, at 1 msec/div and 50 pixels/div that's just 500 samples at a
lazy 50 Ksps rate.

Which makes the oscilloscope totally useless to capture problems. A
good DSO can store loads of samples which where sampled at the highest
possible frequency.

True. I'm not saying that decimation is an optimal or even recommended
approach, just trying to point out to the OP that in and of itself an
ADC that's capable of high sample rates does not mean that it must
always be sampled, stored, and processed at that rate, irrespective of
the display resolution and rate.

It's not aimed at how DSOs from Tek or Agilent approach the problem but
how a reasonable home-brewer might take a first step.

--
Rich Webb Norfolk, VA
.



Relevant Pages

  • Re: Digital Osci and Logic Analyzer
    ... The higher sampling rates of the examples you ... would the o'scope display at 1 usec/div'n? ... time soon) and I have a probe that works properly with the ADC. ... however many samples are collected at whatever resolution ...
    (sci.electronics.design)
  • Re: DSO2250 AC-DC coupling
    ... The apparent change was due to the shift of the waveform on the ... Sounds more and more like they do the offset before the ADC and the ... display limits (this evidenced by manipulating math functions to ... pronounced on channel 2 (the channel used to demonstrate in the ...
    (sci.electronics.design)
  • Re: monitor for macmini mb139ll/a
    ... $150 worth of cables. ... It would plug in, since that monitor uses DVI, but you would get the ... 2560x1600 native resolution of the display. ... You cannot get that resolution out of your Mac Mini. ...
    (comp.sys.mac.hardware.misc)
  • RE: Problem in setting Display resolution from TD
    ... I do not see the display resolution settings from TD ... driver so it means the driver is working fine in Windows XP Prof.. ...
    (microsoft.public.windowsxp.embedded)
  • Re: still no way to do custom aspect ratios?
    ... The good news is that you have found a way to display 4:3 material in normal mode. ... The video that Media ... the PHYSICAL aspect ratio of the television. ... desktop resolution and thinks the resolution in the matching aspect ...
    (microsoft.public.windows.mediacenter)