Re: Scoping small current pulses with Tektronix TM500 series?
- From: Jim Yanik <jyanik@xxxxxxxxx>
- Date: 6 Aug 2006 18:16:05 GMT
"Fred Bartoli"
<fred._canxxxel_this_bartoli@xxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxxx> wrote in
news:44d4d86a$0$13532$626a54ce@xxxxxxxxxxxx:
"Jim Yanik" <jyanik@xxxxxxxxx> a écrit dans le message de
news:Xns9815D1ACF393Ajyanikkuanet@xxxxxxxxxxxxxxxxx
laladong@xxxxxxxxxxx wrote incircuit.
news:1154737386.036397.74930@xxxxxxxxxxxxxxxxxxxxxxxxxxx:
Hi Group,
I am looking for a way to measure the waveform of a current pulse,
on the other of 1-10uA, and with a rise time of ~1us, without
disturbing the 0.5V DC signal that is forced on this node. I was
told this is possible with a combination of Tektronix TM502A (power
module mainframe), AM503B (current probe amplifier module with
digital scale readout), and an oscilloscope. I was wondering if
anyone here has experience with such measurements using this set
up, or may suggest a better method. From reading the AM503B manual
I get the impression that it does not have uA resolution, but then
I could have missed soemthing.
Many thanks.
Dan
Make/insert a small 10 turn wire loop to place the A6302 probe
around;that multiplies the current sensitivity by 10.
You may not disturb the signal enough to matter,it depends on your
IIRC,the highest sensitivity of the AM503/A6302 is 1 ma/div.
Of course,you can also change the display scope's Volts/div setting
to magnify your resolution.(there are limits,though)
And AC couple the scope,or use one with an adjustable DC offset,like
a 7A13/7K combo.
IIRC,that's set at 50mv/div,50ohm inputZ.
AM503 system BW is 50 Mhz,IIRC.
BTW,I used to repair/calibrate AM503 systems at TEK-Orlando.
(before they shut all the field offices down)
Glad you said that.
I just got hold of an AM503 that had been severely "adjusted" to the
point its sensitivity was off by more than 25%.
You have to watch for metal film gain resistors changing value.Then you end
up at the limits for gain pots,which may affect freq.comp.
I went through all the cal procedure, but got a bit confused by the
'adjust until getting good results' for the HF compensation resistors
and cap.
Do you have any good hint about this?
Well,the AM503 amplifier can be "pre-adjusted" without the A6302 probe by
driving a signal into the connector,IIRC,there was instructions on how to
do this in the manual,and how to make the adapter.
One thing is that the A6302 probe is sensitive to being dropped and that
can crack the ferrite head or damage the Hall device and you'll never get
it to compensate properly.
also,some of those tiny adjustable caps can have the slot break loose from
the rotor disc and it looks tlike they're adjusting,but nothing
changes,because the disc itself is stuck and not turning with the metal
'slot'.They can also develop a black silver oxide underneath them and that
insulates the rotor from the metal stud,making it effectively open.
BTW,TEGAM bought the TM500 line from TEK,and were providing parts and
service for AM503. I'm not sure if this is still true though.
www.tegam.com
--
Jim Yanik
jyanik
at
kua.net
.
- References:
- Scoping small current pulses with Tektronix TM500 series?
- From: laladong
- Re: Scoping small current pulses with Tektronix TM500 series?
- From: Jim Yanik
- Re: Scoping small current pulses with Tektronix TM500 series?
- From: Fred Bartoli
- Scoping small current pulses with Tektronix TM500 series?
- Prev by Date: Re: Scoping small current pulses with Tektronix TM500 series?
- Next by Date: Tek TAS250 probe compensation problems
- Previous by thread: Re: Scoping small current pulses with Tektronix TM500 series?
- Next by thread: Re: Scoping small current pulses with Tektronix TM500 series?
- Index(es):
Relevant Pages
|