Re: Measuring signals with spectrum analyzer



JensGraf@xxxxxx a écrit :
Hallo,

I tried to measure the noise of a DSP-output when no signal is applied
to the output. I measured the "rest"-noise with an oszilloscope. The
amplitude of the signal seems to be about 2mV. Then i wanted to
analyse the power spectrum of the noise. I connected the DSP-output to
a spectrum analyzer.
Now the problem is that i get different results when changing the
frequancy span of the spectrum analyzer. When i measure from 10Hz to
3.2KHz, the sprectrum of the measured noise lies at about -100dBVrms.
Increasing the span (10Hz to 25,6KHz) leads to a spectrum that lies at
about -90dBVrms. Compared to the previous measurement it is
approximately 10dB higher. Has anybody an explanation for this effect?
How can i ensure a precise measurement of the noise?


What you probably see (probably because we don't know which SA you have and which measurement you did exactly) is this:
when you increase the frequency span, you also increase the SA IF bandwidth (most SAs have frequency span, IF and video BW coupled in order to make the measurement in a 'reasonable' time).

What you want is a noise power density (nV/rtHz or uV/rtHz) and what the SA gives you is nV or uV. For each analyzed frequency it gives the total noise power that is within the IF BW around the analyzed frequency (hope that's clear enough). As you increase span, IF BW increases and so does the noise power.

To convert to PSD you'll have to check in your manual how the noise BW of your SA is related to IF BW (it may not be identical, but often is), and then divide the noise power by the sqare root of the noise BW.

Then the obtained figures should be identical.

Some analyzers also have the possibility to directly do noise PSD measurements at some spot frequencies during the sweep. Check your manual for this.


--
Thanks,
Fred.
.



Relevant Pages

  • Re: Confidence limits for noise measurement
    ... required to reduce the effect of chance on the measurement result ... The first thing to do is calibrate the instrument against a standard ... white noise, an attempt to measure the noise source by sampling the ...
    (rec.radio.amateur.antenna)
  • Re: How to measure really small voltages
    ... >>measure this small voltage? ... One thing nobody has mentioned is the cross-correlation method. ... can be made AC by a variety of methods, but the additive Johnson noise is ... In the cross-correlation measurement, you take two voltmeters (made e.g. from ...
    (sci.electronics.design)
  • Re: Low noise IF3601/02 dont meet specs
    ... How much noise have you measured ... >> I have read that thickfilm resistors have more noise than ... The question is now about the used noise measurement method. ...
    (sci.electronics.design)
  • Re: [Q] vector signal analyzer, dynamic signal analyzer, carrier noise test set
    ... > I am interested in the phase noise of carrier frequency. ... > I've heard that equipments such a vector signal analyzer, ... spectrum analyzer that would otherwise be obscured by the analyzer's own ...
    (sci.electronics.equipment)
  • Re: Sun noise
    ... I have done two series of 250 measurements of audio noise voltage ... estimate of the uncertainty that I referred to in an earlier post. ... distribution, so the variance is 1/12 of the span. ... and that might also be true of measurement of noise voltage ...
    (rec.radio.amateur.antenna)

Loading