carneyke wrote:

You may have something causing the offset (bad solder connection). The
unit may have been damaged. The pot is for calibrating the 300 mv dc
range. As you have found it is the only adjustment. You either put in
an accurate 300mv or 3Volts and adjust the pot. Everything else falls
in to place (unless it has been damaged). Good Luck.

I took it apart and resoldered the studs for the input leads. I also took both
fuses out and used emery cloth on the contacts.
It now reads about .8 to 1 ohm leads shorted. My other 75 reads about .2 ohms
leads shorted. And yes i swapped the leads to see if it made a difference. (it
did not).

Voltage readings was about .1 volt difference between the two on DC volts. The
meter is not useless, i just cannot use it to reliably test very low value
resistors.

I wonder what could affect the ohms scale and not the DC volts?

Bob




For a digital meter, that 0.1V difference could be significant (unless you're
measuring an unstable source). What voltage range did you check on the meter?
If the lowest range, then I'll bet that's your problem.

--
Dave M
MasonDG44 at comcast dot net (Just substitute the appropriate characters in the
address)

Life is like a roll of toilet paper; the closer to the end, the faster it goes.


.



Relevant Pages

  • Re: Fluke 77 ohms reading cal off
    ... The pot is for calibrating the 300 mv dc ... As you have found it is the only adjustment. ... an accurate 300mv or 3Volts and adjust the pot. ... Voltage readings was about .1 volt difference between the two on DC volts. ...
    (sci.electronics.repair)
  • Re: Fluke 77 ohms reading cal off
    ... As you have found it is the only adjustment. ... an accurate 300mv or 3Volts and adjust the pot. ... I took it apart and resoldered the studs for the input leads. ... Voltage readings was about .1 volt difference between the two on DC volts. ...
    (sci.electronics.repair)
  • Re: 6V Meter on 12V
    ... This old meter has a screw where I can adjust the full scale landing ... That screwdriver adjustment is really a "fine tuning" adjustment to ... A series resistor equal in value to the internal resistance of the ... volts with 12 volts applied to the resistor/meter combination. ...
    (rec.boats.electronics)
  • Re: 50A DC Current Sensing
    ... 50mV meter. ... small range pot on it in the middle of the divider to calibrate ... you have 4.5 volts for full 50Amps, ...
    (sci.electronics.design)
  • Re: Potentiometer of Technics 323 tuner amplifier
    ... I got Technics 323 tuner amp, which is lacking a volume pot. ... balance adjustment on the same knob, ... So might the balance adjustment be done so that the "head" just adjusts ...
    (sci.electronics.repair)