Re: measuring the pulse length of a nanosecond laser



In article <1177689223.353838.195260@xxxxxxxxxxxxxxxxxxxxxxxxxxxx>,
meto <metelek83@xxxxxxxxxxx> wrote:

Hello
I want to measure the pulse length of a nanosecond laser(at 532nm Nd-
Yag) with a photodiode.
I could not find the correct photodiode to measure it.
Our lasers beam length is approximately 4-5 nanoseconds but I don't
know the exact length and I wanna learn it.
Can you suggest me a detector anf manufacturer for measuring it.
If you have another idea to measure the pulse width can you let me
know?
Thanks


This isn't a trivial measurement. After you've converted your light
pulse to an electrical signal with your photodiode, you have to measure
the width of the electrical signal. If its 4ns then to get a good
measurement with an oscilloscope you'll need a 2GHz bandwidth machine.
You might get away with 1GHz, check the risetimes of the instrument.
You might be able to work on the signal with fast discriminators and a
time-to-digital converter (TDC) but setting up without a fast scope is
likely to be a problem.


There is a very clever technique for measuring the widths of nanosecond
light flashes - called the single-photon technique, also known as
"time correlated single photon counting".
It uses devices whose time response is comparable with the light
flash being measured - and samples over many flashes to derive
the shape.

The original paper is

Bollinger L M and Thomas G E 1961 Measurement of the
time dependence of scintillation intensity by a delayed coincidence
method
Rev. Sci. Instrum. 32 1044-50

If you put Bollinger Thomas photon into google or similar you'll find
lots of info.
.



Relevant Pages

  • Re: measuring the pulse length of a nanosecond laser
    ... I want to measure the pulse length of a nanosecond laser(at 532nm Nd- ... I could not find the correct photodiode to measure it. ... This isn't a trivial measurement. ...
    (sci.optics)
  • Re: measuring the pulse length of a nanosecond laser
    ... meto writes: ... I want to measure the pulse length of a nanosecond laser(at 532nm Nd- ... I could not find the correct photodiode to measure it. ...
    (sci.optics)
  • Re: measuring the pulse length of a nanosecond laser
    ... I want to measure the pulse length of a nanosecond laser(at 532nm Nd- ... I could not find the correct photodiode to measure it. ... A fast biased PIN detector like a Thorlabs DET25K should work; ...
    (sci.optics)
  • Re: I Owe Einstein an Apology. Sorry Albert!
    ... >> pulse from A to B to C ... >> That's a measurement, not an assumption. ... If OWLS!= c then OWLS cannot be isotropic. ... > OWLS experiment with two spatially separated and synchronized clocks. ...
    (sci.physics.relativity)
  • Re: How to measure on/off LED timing ?
    ... I need to provide the LED measurement result for compare ... What I should do is to use a oscilloscope and check the current? ... moderately high-speed photodiode; at least 2X faster than the LED ... voltage drop across 50 ohm input impedance. ...
    (sci.electronics.design)