Re: Zeiss objective




"J" <J@.com> wrote in message
news:EutWg.11644$O65.10503@xxxxxxxxxxxxxxxxxxxxxxxxx
I purchased a Zeiss Phase Contrast Microscope and I am interested in
learning more about the objectives.

The Zeiss objectives are labeled: PH1 F-LD. Are these good objectives?

I've seen PH1, PH2, and PH3 objectives. I know PH is the abbreviation for
Phase Contrast. What does the number after the PH mean?

What does the F abbreviation stand for?

Phase contrast was developed by Fritz Zernike who won the Nobel prize in
1953. The numbers on the objective should correspond to numbers on the
condenser. So when you are using a phase contrast 2 objective you will have
2 on your condenser and the condenser will be centered. This is each
manufacturers way or identifying the objective to condenser match. This
also refers to NA. There is no international match criteria but this is the
system nearly universally used. The only exception I can remember was AO
used colors.

Kevin Cunningham
SMS


.



Relevant Pages

  • Re: interference phase for old Nikon S
    ... Dennis is talking about a unique Nikon item which allowed Zernike phase ... contrast to be used without proper phase contrast objectives. ... phase condenser. ...
    (sci.techniques.microscopy)
  • Zeiss objective
    ... I purchased a Zeiss Phase Contrast Microscope and I am interested in ... The Zeiss objectives are labeled: PH1 F-LD. ...
    (sci.techniques.microscopy)
  • Re: Entrance Pupil Microscope Objective, Now what about the condenser?
    ... >> Must the condenser be somehow telecentric also? ... Between the objective and the so called tube lens. ... telecentricity in object space appears to be a property of all ... microscope objectives where as infinity correction refers specifically to ...
    (sci.optics)
  • Re: How to setup Zeiss DIC?
    ... A Carl Zeiss standard microscope, several objectives (normal and phase ... what other parts do I need to make the DIC condenser work ... Zeiss" on the top surface) then you need the Zeiss type III slider ...
    (sci.techniques.microscopy)
  • Re: How to setup Zeiss DIC?
    ... A Carl Zeiss standard microscope, several objectives (normal and phase ... - what other parts do I need to make the DIC condenser work ... leaving DIC prisms in place for BF and PC work. ...
    (sci.techniques.microscopy)